Teledyne T265 Weather Radio User Manual


 
Appendix C
Warranty/Repair Questionnaire
M265E and T265
(P/N 074990000 Rev A, DCN# 6167)
TELEDYNE API CUSTOMER SERVICE
EMAIL: api-customerservice@teledyne.com
PHONE: (858) 657-9800 TOLL FREE: (800) 324-5190 FAX: (858) 657-9816
CUSTOMER: ____________________________________ PHONE: ___________________________________________
CONTACT NAME: _______________________________ FAX NO. ___________________________________________
SITE ADDRESS: ______________________________________________________________________________________
MODEL SERIAL NO.: ____________________________ FIRMWARE REVISION: _______________________________
1. ARE THERE ANY FAILURE MESSAGES? ____________________________________________________________
_____________________________________________________________________________________________________
_____________________________________________________________________________________________________
PLEASE COMPLETE THE FOLLOWING TABLE: (NOTE: DEPENDING ON OPTIONS INSTALLED, NOT ALL TEST
PARAMETERS SHOWN BELOW WILL BE AVAILABLE IN YOUR INSTRUMENT)
*IF OPTION IS INSTALLED
PARAMETER RECORDED VALUE ACCEPTABLE VALUE
RANGE
PPB/PPM 100 PPB TO 2000 PPB
O3 STAB
PPB/PPM < 0.3 PPB (RMS) WITH ZERO
AIR
SAMPLE FLOW
CM
3
500 ± 50
NO FLOW
CM
3
5 ± 2
PMT SIGNAL WITH ZERO
AIR
MV -20 TO 150
PMT SIGNAL AT SPAN GAS
CONC
MV
PPB
0-5000MV
0-20000 PPB
NORM PMT SIGNAL AT
SPAN GAS CONC
MV
PPB
0-5000MV
0-20000PPB
AZERO
MV -20 TO 150
HVPS
V 400 – 900
RCELL TEMP
ºC 50 ± 1
BOX TEMP
ºC AMBIENT + 3 -7ºC
PMT TEMP
ºC 7 ± 2ºC
RCEL PRESS
IN-HG-A 2-10 (constant)
SAMP PRESS
IN-HG-A AMBIENT - 1 ± 1
NO PRESS
IN PSIG 5 ± 0.2
O3 SLOPE
1.0 ± 0.3
O3 OFFSET
-20 TO 150
ETEST
PMT MV 2000 ± 1000
OTEST
PMT MV 2000 ± 1000
Values are in the Signal I/O
REF_4096_MV
MV
4096mv ±2mv and Must be
Stable
REF_GND
MV
0± 0.5 and Must be Stable
07337B DCN6324
C-1